Re: Electricity tends to flow through the path of least resistance by AnalogKid ..... Zapper Support Forum

Date:   6/18/2015 10:45:26 AM ( 10 y ago)
Hits:   3,538
URL:   https://www.curezone.org/forums/fm.asp?i=2260302

0 of 0 (0%) readers agree with this message.  Hide votes     What is this?

1. For some reason I thought you were in Georgia.

2. As I have pointed out before, 60601 does not require transient protection that is superior to that in other standards (the most common being 60950).

60601 requires a lower maximum leakage current compared to 60950. This leakage current is a) a steady-state condition, not related to transients in any way; and b) handled by bypass capacitor selection, not transformer design.

But wait, there's more. Because a 60601 supply has a *higher* transient impedance to earth ground (a consequence of the required lower leakage current), it actually is ***more*** "dangerous" than a standard supply; dangerous in your context, the ability to transfer an input transient to the output.

More of those pesky facts...

ak

 